July 2003 Radiofrequency
Catheter Ablation of Complex Atrial Arrythmias in Patients
with Congenital Heart Disease
Ed.note: In this issue of ‘Current Opinion’,
Dr. Andrew Blaufox provides an update on innovative
3-dimensional mapping systems that are being used to
guide radiofrequency ablation in patients with congenital
heart disease.
Andrew
D. Blaufox, M.D.
Children's Heart Program of South Carolina
Charleston, South Carolina
There has been a tremendous increase
in the number of patients who are reaching adulthood
after having undergone palliative surgery for congenital
heart disease, thus there has been a growing need to
focus attention on the late term morbidity suffered
by these patients. Among the more prevalent, disruptive,
and potentially lethal problems experienced in this
population is intra-atrial reentrant tachycardia (IART).
The cumulative occurrence rate for this arrhythmia approaches
50% at 12 years following palliation procedures, such
as the Fontan or Mustard procedures. Although the incidence
is not as high after surgery for other defects, such
as Atrial Septal Defects or Tetralogy of Fallot, it
is still significant. Because tachycardia cycle lengths
are typically longer in IART than those seen with typical
adult atrial flutter and because patients with IART
often have fairly well preserved AV conduction, there
is a greater chance that these patients will have 1:1
AV conduction with rapid ventricular conduction. This
may lead to more severe symptoms or sudden cardiac death.
Thus, there is a strong impetus to treat IART aggressively.
Pharmacologic therapy has proven to be
ineffective for long term control with recurrence rates
greater than 75% at 5 years. The results being reported
for Fontan revision (total cavopulmonary artery conversion
combined with right-sided maze or Maze-Cox III surgery)
have been very promising with recurrence rates as low
as 12.5%. However, these results are only applicable
to Fontan patients, especially those with significant
hemodynamic issues requiring intervention with a substantial
morbid risk. On the other hand, radiofrequency catheter
ablation of atrial arrhythmias can be safely undertaken
in a wider set of patients with a much lower morbidity.
Although cumulative recurrence rates are approximately
40%, patients who have recurred have improvement in
their symptoms and quality of life. As experience and
knowledge are growing rapidly, current practices are
likely to yield greater success. Advances in mapping
utilizing three dimensional techniques and advances
in lesion formation utilizing catheter tip cooling promise
even greater improvement in ablation results.
These atrial arrhythmias consist of electrical
circuits around barriers to electrical conduction, such
as scars or incision lines, or between areas of electrical
isolation, such as the orifices of the caval veins or
the atrioventricular valve annuli. Before the advent
of advanced mapping techniques, identification of these
structures relied heavily on descriptions from surgical
notes and other conventional mapping techniques. Various
specific diagnostic catheters have been used. Typically,
one or two catheters with 20 electrodes were laid out
in the chamber of interest so that atrial activation
patterns could be carefully analyzed. Areas of slow
conduction were sought. The ability to produce concealed
entrainment and post-pacing intervals within 20 msec
of the tachycardia cycle length indicated that these
areas of slow conduction were critical to the tachycardia
circuit and should be targeted for ablation. Ablation
results utilizing this type of mapping varied from center
to center with reported acute success rates of 70-100%
while recurrence rates were approximately 50% at 2 years
follow-up.
Although there was an appreciation for
the complexity of these atrial circuits, knowledge of
the underlying electrophysiological and anatomical substrate
of the critical protected zones of the reentrant circuit
was limited with conventional mapping techniques. The
use of innovative mapping systems, such as CARTO (Biosense-Webster,
Johnson-Johnson) and ENSITE3000 (Endocardial Solutions)
have greatly facilitated better understanding of the
circuits and have been associated with a greater long-term
outcome after radiofrequency ablation procedures. Each
system operates from a different technology and has
benefits and disadvantages in comparison to the other.
CARTO operates via an electromagnetic
field to provide spatial orientation data for anatomic
reconstruction and catheter positioning without the
need for fluoroscopy. Electrograms are acquired point
by point with a single roving catheter and are retained
with easy ability for review. Activation timing is calculated
in comparison to a reference catheter. Activation is
color coded for display and interpolated between collected
points. In addition, voltage potentials are recorded
with each point, stored, and can be displayed in a similar
fashion as activation timing.
This system has several advantages:
- activation timing that is easily
be seen and updated with numerous points and can readily
be correlated with anatomy,
- catheter position can be easily and
accurately marked during mapping and ablation so that
repositioning can occur more accurately,
- scars, prosthetic material, and anatomic
obstacles can easily be identified and marked as barriers
to electrical conduction,
- voltage maps can provide valuable
information regarding potentially critical parts of
the circuit,
- the completeness of bidirectional
conduction block following radiofrequency applications
can be readily assessed, and
- the electroanatomic data can be obtained
in relatively small or narrow chambers, such as systemic
or pulmonary venous atria in Mustard or Senning patients.
The disadvantages of the system are in
part due to limitations of the technology as well as
the problems with the substrate.
- In order to obtain more detail about
activation, more points need to be collected.
- As data between points is interpolated,
the potential for error is dependent upon the number
of points taken, thus prompting the user to spend
more time gathering as many as hundreds of data points.
- Because many of these patients, particularly
Fontan patients, have multiple circuits, data collection
for one circuit may be interrupted and confused by
initiation of another circuit.
- Time constraints make complete data
collection difficult for nonsustained or hemodynamically
compromising tachycardias.
Nonetheless, CARTO has proven to be a very valuable
tool.
Figures
1 and 2
are CARTO propagation maps of an IART circuit and results
from its ablation in a 5 year old patient s/p Tetralogy
of Fallot repair who had previously failed Amiodarone
as well as Sotalol. Figure
1 demonstrates a counterclockwise IART around an
anterior atriotomy scar through a channel between the
scar and the IVC. An ablation line was placed between
the scar and the IVC as well as between the scar and
an ASD patch. Figure
2 shows conduction block during medial pacing after
completion of the RF line.
ENSITE relies upon an array with approximately
3,000 electrodes that is opened around a fluid-filled
balloon enabling it to capture electrograms that are
virtually projected, via Laplace’s law, onto an
anatomy which was constructed using a single roving
catheter. This system has some very distinct advantages:
- activation is recorded, displayed,
and correlated with anatomy on a single beat so that
extensive amounts of time for data acquisition are
not required allowing one to map multiple circuits,
nonsustained tachycardias, and hemodynamically compromising
tachycardias,
- electrogram amplitude is readily seen
so that inferences about tissue health can easily
be made,
- catheter position can be easily marked
during mapping and ablation, and 4) the completeness
of bidirectional conduction block following radiofrequency
applications can be readily assessed.
Although this system has proven to be
accurate, several disadvantages exist:
- interference from farfield signals
may influence electrogram interpretation,
- the array is thrombogenic and requires
vigilant attention to aggressive anticoagulation,
- scars, incision lines, and other
barriers to conduction are somewhat more difficult
to identify,
- the balloon size limits the use of
this system in smaller chambers seen in smaller patients
or certain anatomies, such as in the Mustard or Senning
patients, and
- the anatomic and spatial accuracy
is limited in very large chambers such as those seen
in some older APC-Fontan patients.
Nevertheless, ENSITE has helped tremendously in dealing
with these arrhythmias.
Figures 3 – 7 demonstrate the use
of ENSITE to delineate the IART circuits and results
from their ablation in a 15-year-old boy with Tricuspid
Atresia s/p APC Fontan. Figure
3 shows tachycardia #1. This tachycardia travels
in a counterclockwise direction: anterior and inferior
then superior and posterior through a gap in the mid
cristae terminalis. Figure
4 shows tachycardia #2. This tachycardia travels
in a counterclockwise direction: starting superior and
posterior then heading anterior by passing below the
cristae terminalis through a channel between the cristae
and the IVC. A line of radiofrequency applications was
made along the posterior aspect of the cristae terminalis
through its gap and down to the IVC. Figure
5 shows unidirectional conduction block with posterior
pacing following ablation. Figure
6 shows that a undirectional gap in the ablation
line exists in the mid cristae region when anterior
pacing is performed. After this is realized, three more
lesions are placed along the anterior aspect of the
cristae in order to close this gap. Figure
7 demonstrates that the gap is closed with these
final lesions.
In addition to these important advances
in mapping, the advent of catheter tip cooling promises
to allow physicians to create more complete and transmural
lines of conduction block. Cooling at the catheter tip
actively with irrigation or passively with larger tip
surface areas allows for lower impedance and greater
energy transfer and heating into the tissue. Initial
reports indicate that acute success is augmented by
catheter tip cooling in adults with atrial flutter as
well as patients with congenital heart disease and IART.
While management strategies for patients
with IART should be individualized according to patient
need, the recent advances in radiofrequency catheter
ablation make it a very attractive option for many of
these patients. |